I just also like to point out that the point of adding Potassium Hexacyanoferrate is to form a complex ion with the iron present in the solution. This is a bit beyond the scope of year 12 I think, or just not a major part. Basically, it forms strong polar bonds around the iron ion, much like when salts ionise in water, except your Fe ions are surrounded by (complex) ligands. That's just for your.. erm.. general knowledge haha
But when used in colorimetry, complex ions are preferred because they often give off an intense distinctive colour. In your prac, you are adding the Hexacyanoferrate to produce the purple colour, and in theory if you prepare your standards correctly, it will not interfere with your final results since the graph is all 'relative' to your reference cell.
You can consider the Hexacyanoferrate added as an impurity / contamination if you want, so if its present in all your standards, in the same amounts, your graph will be adjusted to it. A common example is that of a calibration curve not passing through the origin, because the element you are testing for is present in the solvent (eg heavy metals can be found in water). You can either adjust your equipment, or simply have a calibration curve that begins above the origin.
I'm only guessing at the method you intend to use for your sac, so I may be completely wrong but hope it makes sense.
Good Luck with your sac!